I reran a vendor's method for identity and purity of a polypeptide. Invendor's method, they use Agilent HPLC. However, I used Perkin Elmer HPLC. I follow the vendor's method exactly: the gradient, column type, column temperature, injection volume,etc. My system suit failed. I think the failure maybe due to the gradient since different instrument has different dwell volume. So, I went on and measured the dwell volume of my instrument. I don't know how to adjust the gradient if the dwell volume is known. Can you please help! Thank you,
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
The dwell time (dwell volume divided by flow rate) is effectively an isocratiddc hold at the beginning of the gradient. All you need to do is to purposely add an isocratic hold or a delay to injection to make the total delay the same on both instruments. That said, there may be other factors such as mixing volume, proportioning accuracy, or column chemistry involved.
Knowing what part of system suitability failed might give us more to go on.
Knowing what part of system suitability failed might give us more to go on.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 273
- Joined: Mon Aug 04, 2014 8:13 am
- Location: UK
Hello
If you have 2 different pumps (quat vs. binary), different tubing the total system volume will be different. So in my opinion the best option is to adjust gradient manually just to achieve correct parameters (resolution between peaks, retention time range).
Regards
Tomasz Kubowicz
If you have 2 different pumps (quat vs. binary), different tubing the total system volume will be different. So in my opinion the best option is to adjust gradient manually just to achieve correct parameters (resolution between peaks, retention time range).
Regards
Tomasz Kubowicz
-
- Posts: 252
- Joined: Sat Nov 07, 2009 6:27 pm
What is the QA used by the vendor? Metric for performance?
-
- Posts: 6
- Joined: Wed Nov 16, 2016 4:35 pm
Is anyone aware how to create an injection delay to an Agilent 1260 running chemstation? The injector program function eludes me. Much more familiar with MassLynx.
-
- Posts: 686
- Joined: Mon Apr 29, 2013 4:37 pm
- Location: 39° 44' 44" N / 75° 32' 48" W
MattM
6 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science