- admin
- Site Admin
I have developed a combined method for the analysis of the antioxidants BHA and Propyl Gallate in pig feed samples, with 65/35 Meoh/THF as the extraction solvent the Propyl Gallate peak will occaisonaly degrade rapidly. Has anybody got any suggestions for improving the stability of the solution (e.g. a more sacrificial antixidant) or a suitable glassware cleaning procedure.
By Consumer Products Guy on Thursday, May 27, 2004 - 07:47 am:
Dissolve in DMF, make trimethylsilyl derivatives (to make stable), GC on non-polar capillary column.
By Anonymous on Thursday, May 27, 2004 - 07:55 am:
Chris,
I wonder if there might be some trace peroxides in your THF. These might degrade the anitoxidants pretty rapidly also. Can you get just as good extraction efficiency w/o the THF? MIght be worth an experiment or 2.
Mark
By Chris Allen on Thursday, May 27, 2004 - 11:51 am:
I am currently using HPLC grade THF which has a maximum peroxide content of 0.05%, but there is the option of using AR grade which is stabilised and has a maximum of 0.001% peroxides. The degradation does not appear across a whole batch of extraction solvent some solutions will degrade within a couple of hours whilst others are stable overnight. Any suggestions on glassware or vial contamination?
By Anonymous on Friday, May 28, 2004 - 11:43 am:
THF stabilized with BHT as an extraction solvent could improve your stability. BHT is an antioxidant it self and could be consumed preferentially in the solution.