DJ wrote:Connie Parker wrote:
Tom I dont think you did misunderstand my initial question I guess I was just asking the wrong question.
So how do I determine if my analyte is being retained (interacting with the stationary phase efficeintly) on the column when running a single gradient screen? i.e. that k* is between 1 and 10 if we cannot measure k*?
Do I just estimate S as being ~4 (or 5 as I've also seen in the literature) as DJ suggests and use the equation to find K*?
How do you determine if your analyte is being retained? Shoot some uracil on your column. If you compound has the same retention time, it's not being retained. (or did I mess the question )
DJ I guess the emphasis in my statement should have been on the word "efficiently". Your right of course anything retained after uracil on a normal reversed phase column will be “retained” but k* as I understand it is used as an indication of how efficient the interaction of your analyte with the stationary phase is. k* as with k’ in isocratic runs should be between 2 and 10 although I suspect if you get a nice peak with a half decent retention most folks would run with it regardless of what k* is.