I am running HPLC samples use the same method and same sequence except I prepare a new mobile phase A which is 0.1% TFA in water. There appeared 3 huge peak in my blanks .
Another analyst reprepare the mobile phase, we changed mobliphase B and C ( ACN and IPA) and cleaned the system and aloso change another used column. but there is still something on my baseline. I realzied it might cause by water, so I used bottled HPLC water. the base line are not that bad, the noise peak decreased from 0.02 to 0.005 and they are not 3 peaks anymore but they changed to one with a small negtive peak there. Right now I am using good water, ACN and IPA try to clean my system, what else I can do?
how can I post picture?
By the way, the method was running all days long for few months and nevery has such problem. it is gradient from 90% TFA to 70% ACN with 5%IPA in 3.5 minutes and stay there for 1.5 minutes and back to 90% TFA . the column is a 10cm column and the detect at 278nm.
those peaks shows up at 2.5-3 minutes.
This is one of the more common questions that gets asked on the Forum. Watch this video for a discussion of what's going on and a diagnostic test to confirm the cause of the problem (you will have to register separately). http://www.lcresources.com/more_resourc ... hp?f=3&t=5
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
thank you very much. they are very helpful. I think the problem comes from the water. But how can I get fird of it. It used to be very good baseline. We used to use millQ system and we changed erverything we can change. And we changed the water to HPLC grade water, the baseline is better but not enough. And based on the website you posted, I double check, the problem comes from the solvent A. I plan to by deion water instead, hope it helps.
You might also try the "Empore" filtration trick:
www.chromforum.org/viewtopic.php?f=1&t=130
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry