I have been using a Kinetex PFP column, 100 mm x 2.1 mm, 2.6 uM, 100A (it has an effective filter pore size of 0.2 uM) and an Opti-Solv EXP pre-column filter (with titanium Hybrid ferrule) that also has a 0.2 uM pore size. Is there any reason that these two should not be compatible? I had an issue with my new column becoming clogged, but the guard column was not clogged. I can't imagine how the sample could pass through the guard column and not the actual column. I am using filtered plasma samples acidified with TFA. I have used these samples with an identical setup in the past with no problems. I am also filtering the plasma and precipitating proteins.

The company that makes the column claimed it was likely because I was not using their guard column, but this one has been fine in the past, and it appears to be compatible based on the specs. I am awaiting to hear from the company that makes the guard column.

Any ideas? Thank you in advance!!!
Hi,

I have the exact same column and it clogged a while ago too. But I try to wash it according to the manual and now it worked just fine.

I don't exactly know what caused the problem, but it was new (usually they ship in ACN) and I did not wash column with water before I run my phosphate buffer mobile phase. So it may be the phosphate precipitate due to the high buffer conccentration? (Anyway, is your mobile phase phosphate buffer?)

Since your guard column was used, so the precipitation is occurred in the kinetex column?

I suggest a wash/flush can regenerate your column.
2 posts Page 1 of 1

Who is online

In total there are 410 users online :: 2 registered, 0 hidden and 408 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 408 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry