If the amount of analyte is constant in the matrix you're using for preparing standards, you might be able to use the method of standard addition to initially determine the background concentration, then take that into account when preparing your standards for your calibration curve.
For example I have an ICP method for iron that has 0.3ppm of iron in my matrix, which I determined by standard addition and confirmed with outside analysis. I calibrate using the straight matrix sample and a matrix sample with 50ppm spiked - so my curve is a 0.3ppm point and a 50.3ppm point.
Whether this is acceptable would depend, among other things, on whether you can accurately determine the low levels of analyte in your matrix, what kind of regulatory environment you are in, and any internal SOPs you have to follow.