I inherited a Varian ProStar binary setup (PDA, Prep heads, vacuum degasser) and I seem to be having some trouble with analytical injections.
Currently running a waters C18 4.3mm analytical column with a SecurityGuard cartridge, running MeOH/H2O mixes.
The system will run fine if left to pump by itself, blank MeOH injections into either pure MeOH or anywhere down to 10:90 MeOH:H2O is fine, a very small peak due to solvent composition change or small impurity, but this is okay for our application.
The problem is that when I take my analyte, dissove it in either 100% meoh or 50:50 MeOH:H2O, and degass the solvent (sonicator), injections into the system cause lots of gas bubbles to come out of the detector.
I thought maybe the rheodyne was leaking but i would have seen this if my blank injections were gassing out. I carefully prepare my syringes (gas-tight, fill past injection volume, vertically purge out headspace) but still I only get gassing out when I inject a real sample.
Chromatogram is a single signal at about 5 minutes, a very broad one, with high frequency noise about 50% of the vertical intensity of the main peak height. For those of you familiar with NMR it looks like a septet essentially. Corresponding to the high frequency transients are loads of gas bubbles out of the PDA line, much much more than I could have possibly injected into the rheodyne. Tried purging/washing out the injector loop, didn't help.
Not sure where to go from here - anyone have any idea why my sample is gassing out?
edit: If it helps, C18 TLC shows three resolved spots, of which the sample is a "click" product - aliphatic alkyne and azide --> 1,2,3-triazole. I suspect 2 spots are starting material and 3rd one is product.
Regards,
Seb