acarbose hplc condition

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
hi

i am anew friend in this php .if any one can help me to fined the suitable hplc method for acarbose .i havenot NH2 column if there is another column i can use it instead of NH2 column
dr smms

Don't bother with a NH2 column. You only will have trouble.

Look for the BEH Amide column at the Waters website.

I also would say that NH2 column is not the best choice in terms of reproducibility.
If you google for Acarbose you will find a lot of published methods. From Anion exchanger, to special sugar columns, affinity columns, base deactivated RP-18 columns, standard RP-18 columns to NH2 columns. My recommendation is to use a well endcapped RP-18 of the new generation columns. I would go for a so called polymeric bonded column, means trifunctional bonded, or difuncional bonded with a difunctional bonded endcapping, where you can use up to 0% organic modifier in mobile phase, or just try HILIC. Please take into consideration to use a volatile buffer, if you want to use MS detection later on. Anyhow, if you have a virgin RP-18 column in your lab, endcapped, you can use it. Good luck
Gerhard Kratz, Kratz_Gerhard@web.de

you could try a carboPac PA1 dionex column for IC they have many applications for sugars
or a Sugar-D cosmsil column with 70% ACN/30% water

You can analyze this compound using HILIC column. Here is application for other sugars. In case of higher molecular weight sugars you need to use less ACN (60-65% for your molecule):
http://www.sielc.com/pdf/Sugars%20on%20 ... column.pdf

Below is separation for voglibose and acabarose on Unison UK-Amino:

http://www.imtaktusa.com/site_media/fil ... TI402E.pdf
Hello membres,
I communicate to you the analytical method of Acarbose dosing in a finished product.
The chromatographic conditions for the determination of AI in the FP are the same as in the monograph PH EUR for this AI. However, applying this method in our laboratories, we find a very unstable TR: Tr said 16 minutes, after some injections he shifts and even gets to 9 minutes (! Sometimes less). the column used is: Hypersil APS-2 250 mm × 4 mm, 120 am-5. µm.
Thank you tell us if he has any special precautions for this column.
The APS-2 column is an amino bonded-phase column. If you read the earlier posts in the thread, you will see that experts (including the late and much-missed Uwe Neue) do not recommend that type of column. I would suggest you check with Thermo to see if they can suggest a specific alternative.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
8 posts Page 1 of 1

Who is online

In total there are 396 users online :: 1 registered, 0 hidden and 395 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 395 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry