- Posts: 3
- Joined: Wed May 31, 2023 9:28 pm
- Location: United States
Noisy baseline in DAD
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
I am fairly new to liquid chromatography and have been trying to set up the Agilent 1260 for a new method. Prior to setting up the Poroshell 120 EC-C18 column for sodium benzoate analysis, the column was flushed in its appropriate solvent for long term storage. The old mobile phase bottles were switched out for the new bottles of buffer (7.8 mM ammonium acetate buffer in acetonitrile pH 4.2 (72:28) and (15:85), degassed) for the new method. The system was flushed overnight to remove any salt buildup from the previous method with 10% isopropanol (column compartment heated to 60C), then rinsed with water. With the Poroshell column placed on the HPLC and the column temp off, the system was purged with the new mobile phase for at least an hour, then the purge valve was closed. The column was equilibrated with the new buffer, but the DAD baseline is still noisy (~5 mAU difference). Can anyone suggest what else I can do to reduce the baseline? I can provide more details if needed.
-
- Posts: 127
- Joined: Tue Dec 05, 2006 3:25 pm
- Location: Leipzig, Germany
Is this random noise or do you see a regular pattern? Could be the pump then...
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
My response is not directly related to the detector, but that seems like a complicated mobile phase for detection of sodium benzoate (as benzoic acid). I think if I was doing such assay, I'd simply use 0.5% phosphoric acid/water and ACN (or methanol).
What wavelength are you using?
What wavelength are you using?
-
- Posts: 3
- Joined: Wed May 31, 2023 9:28 pm
- Location: United States
H.Thomas wrote:
Is this random noise or do you see a regular pattern? Could be the pump then...
It appears to be random noise. The scientist who worked on this had a clear baseline in their analysis. If it is a pump issue, how would I be able to tell?
-
- Posts: 3
- Joined: Wed May 31, 2023 9:28 pm
- Location: United States
Consumer Products Guy wrote:
My response is not directly related to the detector, but that seems like a complicated mobile phase for detection of sodium benzoate (as benzoic acid). I think if I was doing such assay, I'd simply use 0.5% phosphoric acid/water and ACN (or methanol).
What wavelength are you using?
This method was worked on by a different scientist before they left, and they had very clean baseline and clear separation in their analysis. The wavelength is set at 225 +/- 4nm, reference 360 +/- 20nm.
-
- Posts: 174
- Joined: Sun Jan 20, 2019 4:58 pm
tidyup8460 wrote:
ammonium acetate buffer... wavelength is set at 225 +/- 4nm, reference 360 +/- 20nm
Wrong combination. Acetate absorbs well at 225 nm; hence, the signal is noisy.
tidyup8460 wrote:
This method was worked on by a different scientist before they left, and they had very clean baseline
Doubtfully.
6 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science