Misinjections
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
One of our HPLCs (Agilent) appears to be experiencing occasional misinjection issues, as evidenced by no or very little response at the analyte retention time. (Most of the misinjections are with verified analytical standards, so we know how much should be there.) This seems to be most prevalent at the first or second injection of the day, and after that, it can run for many injections with no problem, even with a sample it has previously misinjected. Then the next time we run something, it happens again. I've kind of inherited our HPLC maintenance from a departed co-worker,and I don't have a lot of experience in troubleshooting thing such as autosamplers. Any ideas of where to start? Thanks.
-
- Posts: 318
- Joined: Sun Jan 01, 2012 10:21 am
- Location: Turkiye
Rotor seal of the injector may need replacement.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science