I have inherited HPLC performance qualification duties from a coworker who recently left the company. I got very brief training before he left. I should note here that I am a biologist by education and training, not an analytical chemist. The procedure I was shown calls for 5 standards of butyl benzoate, ranging from 1.0 mg/L to 10.0 mg/L, prepared in HPLC-grade water, from an acetonitrile (ACN) stock. We also include a butyl benzoate QC sample at 5.0 mg/L, prepared the same way. I was having trouble getting consistent results until I changed from preparing the samples in water to preparing in 20:80 ACN:water. Still, I seem to be getting some unexpected variability, particularly at the higher concentrations (7.5 mg/L and 10.0 mg/L). My questions are: Are there stability/solubility issues with butyl benzoate that I should be aware of? Should I be preparing these samples immediately before analysis? Should I use a higher solvent concentration? Should I be using a different concentration range? Any general hints on this procedure?
-
- Posts: 2071
- Joined: Tue Aug 17, 2004 7:59 pm
- Location: 38.590967, -90.213236
Compare peak shapes and retention times - are the more concentrated standards eluting earlier and asymmetrically compared to the lower concentration injections?
If so, you may be overloading the column or detector.
If so, you may be overloading the column or detector.
Thanks,
DR
DR
-
- Posts: 15
- Joined: Thu Jun 06, 2013 12:18 pm
DR - Thanks for the reply. Retention times are in excellent agreement, as are peak widths (two criteria for passing qualification). Peak shapes are good also. I should have explained better in my original post, but when I said I get unexpected variability, I mean that linearity falls apart at the higher concentrations. However, if I make them fresh, or get new samples from refrigerated stocks, it all seems to come into line. Hence my question about stability or solubility. But if butyl benzoate is so sensitive to stability or solubility issues, why use it for a standard? It seems to be referenced in many discussions of performance or operational qualification testing.
-
- Posts: 1
- Joined: Mon Oct 16, 2023 9:48 am
- Location: USA
JHC1953 wrote:
DR - Thanks for the reply. Retention times are in excellent agreement, as are peak widths (two criteria for passing qualification). Peak shapes are good also. I should have explained better in my original post, but when I said I get unexpected variability, I mean that linearity falls apart at the higher concentrations. However, if I make them fresh, or get new samples from refrigerated stocks, it all seems to come into line. Hence my question about stability or solubility. But if butyl benzoate is so sensitive to stability or solubility issues, why use it for a standard? It seems to be referenced in many discussions of performance or operational qualification testing.fnaf games
This is useful for me.
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
JHC1953 wrote:
I have inherited HPLC performance qualification duties
Exactly what is the purpose of such performance qualification? Internal, or to serve as "proof" in event of an FDA audit????
After we had validated a test procedure as per cGMP, our management required a yearly Service/Performance Verification by an outside vendor (guy with a tie) and proper paperwork; we never did any in-between performance verifications unless after a repair (which increasingly had to be done by a guy with a tie, no matter how simple.
JHC1953 wrote:
The procedure I was shown calls for 5 standards of butyl benzoate, ranging from 1.0 mg/L to 10.0 mg/L, prepared in HPLC-grade water, from an acetonitrile (ACN) stock.
I cannot fathom why a butyl benzoate standard would be prepared using water instead of an organic solvent. That's certainly not what I would've considered...
-
- Posts: 3
- Joined: Sun Oct 22, 2023 2:09 am
- Location: USA
I wonder what is the better standard to do the HPLC validation testing ? and what kind of validation tests are needed ?
-
- Posts: 3510
- Joined: Mon Aug 30, 2004 11:43 pm
Arc009 wrote:
I wonder what is the better standard to do the HPLC validation testing ? and what kind of validation tests are needed ?
When I was working, I seem to remember Agilent using caffeine standards for HPLC qualifications.
-
- Posts: 174
- Joined: Sun Jan 20, 2019 4:58 pm
PA/PH/OMCL (11) 04 R6
QUALIFICATION OF EQUIPMENT
ANNEX 1: QUALIFICATION OF LIQUID CHROMATOGRAPHY EQUIPMENT
https://www.edqm.eu/documents/52006/128 ... 8491788491
QUALIFICATION OF EQUIPMENT
ANNEX 1: QUALIFICATION OF LIQUID CHROMATOGRAPHY EQUIPMENT
https://www.edqm.eu/documents/52006/128 ... 8491788491
8 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science