Hello everybody,
I need to analyse a sugar mixture (fructose, glucose, isomaltose, sucrose) in water by using a HILIC HPLC column but I cannot separate fructose and glucose because their peaks overlap. Are there any ideas what I could try? I use the following settings:
mobile phase: acetonitrile + water (80:20)
flow: 1 mL/min (I've already tried 0.2 mL/min and 0.5 mL/min without success)
injection volume: 1 µL (samples were diluted 1:1 with mobile phase)
column: hypersil gold amino from thermo fisher
oven temperature: 35 °C
RI detector with 35 °C; 0.5 s rise time; 5 Hz data collection rate

thank you :-)
Seems like you are running this example?

https://appslab.thermofisher.com/App/19 ... -detection

Do your standards prepared in mobile phase retain? What is your sample diluent, you might be injecting too strong a solvent.

edit: I see your sample is in water, try diluting it 1:4 sample:acetonitrile to match your mobile phase conditions for improved retention, although your injection is small that it might work in your strong diluent.
As already mentioned, try to match the mobile phase conditions with your samples. So dilute it directly with acetonitrile, I would even go to 1:5 as to give 80% ACN. If peaks go too low now, you may be able to inject more.

Also adjust the needle wash solvent to high organic (if the system is used also for RP-mode, this can be overlooked)

Another idea would be to add about 20 mM ammoinum formate or acetate to the mobile phase.

If you can also play with the mobile phase composition, try 15:85 or something in between.
Also check the peak width by injecting a single standard. Calculate the plate number and compare to the specification for the column. If the column is old/damaged, or if there is too much extra-column volume, the peaks will be wider than they should be (= low plate number) and they will overlap even if chemically separated.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
copperfoils wrote:
Seems like you are running this example?

https://appslab.thermofisher.com/App/19 ... -detection

--> Yes, almost. In my system is isomaltose and not maltose; and I have no lactose.

Do your standards prepared in mobile phase retain? What is your sample diluent, you might be injecting too strong a solvent.

--> The sample diluent is my mobile phase, so acetonitrile and water (80:20).

edit: I see your sample is in water, try diluting it 1:4 sample:acetonitrile to match your mobile phase conditions for improved retention, although your injection is small that it might work in your strong diluent.


--> I will try diluting 1:4. I hope it will work. Thank you for your advice :)
Hollow wrote:
As already mentioned, try to match the mobile phase conditions with your samples. So dilute it directly with acetonitrile, I would even go to 1:5 as to give 80% ACN. If peaks go too low now, you may be able to inject more.

Also adjust the needle wash solvent to high organic (if the system is used also for RP-mode, this can be overlooked)

Another idea would be to add about 20 mM ammoinum formate or acetate to the mobile phase.

If you can also play with the mobile phase composition, try 15:85 or something in between.


Thank you very much. I will try all of your tips as well :-)
tom jupille wrote:
Also check the peak width by injecting a single standard. Calculate the plate number and compare to the specification for the column. If the column is old/damaged, or if there is too much extra-column volume, the peaks will be wider than they should be (= low plate number) and they will overlap even if chemically separated.


A very good hint! Thank you very much! I will check that out as well :-)
7 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: John Guajardo and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry