Hello,

which is the acceptance criterion of the peak purity? Is a value of 0.95 acceptable? Is there any official guideline about this?

thank you
"Peak purity" (spectral) is a software/algorithm-dependent quantity. No official limits exist.
See also
Empower Peak Purity and Absorbance
viewtopic.php?f=1&t=110344&p=413124&hilit=peak+purity#p413124
No official "specs" or guidelines exist because the results can be tweaked by the operator to reveal any "purity" value you want. If you understand what the software does and what is assumed in using it, then you will understand how any purity values obtained from the software are marketing nonsense.

Here is a link to an authoritative article on the topic.

    Peak Purity Determination by HPLC Diode Array Chromatography Software: Limitations and Uses
; https://hplctips.blogspot.com/2016/12/peak-purity-determination-by-hplc-diode.html
vmu wrote:
"Peak purity" (spectral) is a software/algorithm-dependent quantity. No official limits exist.
See also
Empower Peak Purity and Absorbance
https://www.chromforum.org/viewtopic.ph ... ak+purity/geometry dash bloodbath

An unreliable method of determining whether the separation method is sufficiently selective is to perform a spectral peak purity check. The absence of hidden peaks of impurities beneath the primary peak cannot be demonstrated by this purity check. The absence of any discernible spectral differences between the various components of the main peak (these differences are sought in the peak purity check) may be caused by low concentrations of impurities and the similarity of their spectra to the spectrum of the parent substance. I think this answer to the above question.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry