HPLC separation difficulties

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello, I was radioiodinating insulin, which had previously been done 20 or so times in a lab by myself or others. The final step is HPLC purification. It uses a 60-minute time frame, and the retention period is typically between 35 and 45 minutes (varies between runs slightly). Anyway, this week, the separation of peaks remained the same although my retention time decreased from 35 to 15 minutes. It simply all eluted too soon.

I initially believed it to be poor column equilibration, but it was replicated twice more with the same RT (at 15 minutes). Any suggestions as to the potential cause? Is there any way the column has gone bad?
The two most likely reasons for the retention time to be too early are:
(1) Your mobile phase composition has changed (IOW: It is not what you think it is due to programming, hardware defect, evaporation etc) OR (2) your flow rate has changed (always measure it using a timer and graduated cylinder. Do not rely on the instrument's displayed values).
Columns can and do go "bad", but this is one of the reasons why we dedicate a specific standard (and method) for each column so we can always go back to verify the column's performance (Btw: NEVER use one of the manufacture's "QC" check standards as your column performance std (use a real sample with strong K prime which is related to your actual analysis). These Qc check samples are often worthless for evaluating a column's true performance. They are chosen to show high plate counts and quick QC results during column packing.
Paxton00 wrote:
Hello, I was radioiodinating insulin, which had previously been done 20 or so times in a lab by myself or others. The final step is HPLC purification. It uses a 60-minute time frame, and the retention period is typically between 35 and 45 minutes (varies between runs slightly). Anyway, this week, the separation of peaks remained the same although my retention time decreased from 35 to 15 minutes. It simply all eluted too soon.

I initially believed it to be poor column equilibration, but it was replicated twice more with the same RT (at 15 minutes). Any suggestions as to the potential cause? Is there any way the column has gone bad?


You might have a partial collapse of your stationary phase (assuming it is C18)
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry