I've run into an autosampler problem setting up the PFAS method this year. It seems as though the autosampler doesn't aspirate 1ul samples WHEN a high pressure (aka thru column, pressure @4Kpsi) method is used. It seems to aspirate small samples OK when a Flow Inject (No column, pressure @100psi) method is used. (A 10ul sample seems to give me decent MRM peaks of my standards for my PFAS method but going to that big of a sample size is going to be problematic with actual samples I'm afraid)
What could be going on?
Below is some data of the mass difference after taking sets of TEN injections of water with sample size and conditions described:
1ul - through column (@4500psi) - mass diff = 0.0010g
1ul - bypass column (200psi) - mass diff = 0.0127g
then repeat
1ul - through column (@4500psi) - mass diff = 0.0036g
1ul - bypass column (200psi) - mass diff = 0.0126g
then
10ul - through column (@4500psi) - mass diff = 0.0891g