KH2PO4 with CH3CN gradient how high can I go

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I have 2 app notes one from Agilent one from Phenomenex using KH2PO4 water and an acetonitrile gradient which goes very high. The Agilent one goes to 80% acetonitrile and the Phenomenex one goes all the way to 100% Acetonitrile (though since the KH2PO4 is in A there is none being put in at 100% B).

Is it ok to go that high acetonitrile with even relative low amounts of phosphate 20mM or so? I thought going above 60% acetonitrile risks precipitation in the system.
Whoever compiled those Agilent and Phenomenex instructions must not have spent much time at the bench actually doing HPLC. At ~ 75% acetonitrile, a saturated solution of KH2PO4 is about 5 mM. A solution to this problem is to prepare triethylammonium phosphate (TEAP), which has an appreciably higher degree of solubility in primarily organic solvents, or alternatively to use the sodium salt of methylphosphonic acid. In either case, we prepare our salts by adding the base to a solution of the acid (of known concentration) until reaching the pH we want.

Be advised that if your analyte is charged, then different counterions will give different retention times and selectivity in reversed-phase and HILIC. See my paper from 2008 (introducing ERLIC) and my paper from 2018 in which this subject is studied at length with regard to HILIC.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
That's what I thought.

here are the two apps for artificial sweeteners
https://www.gimitec.com/file/5989-8961EN.pdf
https://www.phenomenex.com/Application/Detail/13339

Some other related app notes just use 0.1% TFA so I will try that first.

From: MODERN HPLC FOR PRACTICING SCIENTISTS by Michael W. Dong

Beware of the possibility for buffer precipitation when mixing it with organic solvents such as acetonitrile (ACN)—the upper limit should be 70–80% ACN. Caution particularly should be exercised in blending ACN with phosphate buffers. Limit the concentration of the phosphate buffer to <15 mM to prevent precipitation, which can cause check valve malfunctions.
3 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Hollow and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry