UPLC-CAD HELP
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Can anyone explain what the current on the graph display on the actual screen of the CAD measures? Why does this current measured in the same units (pA) not match up with the signal the software (such as Empower) picks up?
-
- Posts: 88
- Joined: Mon May 16, 2016 11:50 am
Cad is measured in pA picoamps
Empower uses mv millivolts
The conversion is 10^6
Two different manufacturers, they can choose whatever response units they want.
Empower uses mv millivolts
The conversion is 10^6
Two different manufacturers, they can choose whatever response units they want.
2 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science