While verifying a method for assay & purity of two coumpounds, we obtained bad accuracy results, with recoveries too high at 80% (>102%) and too low at 120% (~95%). Calibration is performed at 100% on the same material, so there is no matrix effect.
(The analysis is performed on a CSH C18 with gradient elution of water/actonitrile with 0.1% TFA, UV detection at 215/220 nm)
I suspected peak saturation, but the peak height is < 1AU and no flattening is observed at apex. I tried to reduce sample concentration (3X dil), but obtained the same results.
We couldn't find a proper root-cause for this, until we looked at peak purity: it appears the main peak is saturated on some portions of the spectrum, but not at the analysis wavelength or within UV bandwidth (2-4 nm).
We perform the analysis at 215/220nm and observe saturation below 205/210 nm (depending of the compound). It takes a dil 10X to avoid saturation.
Could this explain the results we obtained for accuracy ?
We have a second max in UV spectrum at 260 nm that is much lower in absorption, and when the analysis is performed at this wavelength, there is no accuracy issue, so I would say yes, but I don't really understand why...
I also observed that the current wavelengths are in a "slope" of the spectrum, which I think is also problematic, although it doesn't explain why accuracy results are systematically skewed...
Does anyone have insight on this problem ?