Dear all,

I am wondering if anyone has a similar problem with their RI detectors occasionally. This problem happened to me on a Shimadzu and Waters RI detector. I am grateful if anyone can give me some pointers. Thanks!

Problem:
The baseline drift down slowly and steadily, and then suddenly rise back to the starting level.

Image
URL: https://ibb.co/Zz29NBB

Image
URL: https://ibb.co/0tb89J0

Condition: HPX-87H column, 60oC, 0.5mM H2SO4 0.6mL/min, Waters RI detector.
The mobile phase is properly degassed (sonicator + vacuum for about 1 hr)
Anyhow, that looks like gas bubbles in the reference cell. Connect a 30cm small ID capillary ´to the outlet of your RI detector. That will help.
Gerhard Kratz, Kratz_Gerhard@web.de
Hi Gerhard,

Thanks for the tip. As a RI detector novice, I worry this could put backpressure to the flow cell and break it.
Can you explain a bit more why this backpressure can be helpful? If I get asked by my technical officer, at least I want to be able to explain it. Thanks!
It prevents out-gassing and bubble formation in the flow cell.

Yes , the outlet tubing for RI detector must be carefully chosen , to prevent much backpressure , 1 meter of 0.5mm ID tubing recommended.
4 posts Page 1 of 1

Who is online

In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: No registered users and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry