Retention Time Stamp not Appearing

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I am Calibrating and have run my CRM. There are 11 nice peaks with good shape. Only 10 of them have Retention Time Stamp. The last one has no time stamp. and does not appear in the Cal Table either. How do I add it ??


Image

Thank you !
Solved ! Integration events END was at 7 min and this peak was at 7.1min. Adjusted event to end at 7.25 and baseline, RT and area appeared