No, the G1379A/B internal vacuum tubing kits ARE NOT compatible with and Do NOT fit the model
G1322A degassers (not "G1322B", no such P/N). Only the tubing kits built for the model G1322A will work with the G1322A degassers. Agilent still sells the model G1322A degasser (best one they make). You can still buy the tubing kits direct from Agilent Technologies in Germany. Just order one tubing kit using
Agilent's Part Number, G1322-60002. *Replace the tubing at least every 5-years or anytime a major component part is replaced (i.e. for G1322A replace tubing when a: vac pump, vac valve, chamber and/or sensor is replaced).
Caution, NEVER, use automotive or other non-compatible tubing inside the degasser modules. That type of tubing will contaminate the vacuum flow path of the degasser AND your column and mobile phase with contaminated material (as the vacuum system and mobile phase flow paths SHARE the same permeable paths inside the degasser channels. That cheap tubing leaches plasticizes and contaminates. It is not "
downstream" of anything). The vacuum tubing used inside a vacuum degasser must be of high-purity and chemically compatible with the HPLC solvents used in the mobile phase and the vacuum system (that is why special tubing is made for these modules).
For customers in the USA, they can purchase higher quality vacuum degasser tubing kits for many brands and models of HPLC vacuum degassers from Chiralizer Services for prices far below what Agilent charges. * I work for Chiralizer Services and in addition to providing HPLC consulting, advanced column switching and selection systems for HPLC and LC-MS systems, we have been professionally servicing all brands and models of HPLC vacuum degassers for many of the manufacturers and end-users since ~ 1993 [
https://www.chiralizer.com/hplc-degasser-repair.html ]