Separation of poistional isomers 2,4,5-Trifluorophenyl aceti

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
Hellow Experts,

I am working on separation of 2,4,5-Trifluorophenyl acetic acid and its positional isomers on RP HPLC. Any body have some of literature regarding to separation of positional isomers or some valuable suggestion so that I can get proper separatioin

Best regards
Praveen
Praveen, you can separate positional isomers with our mixed-mode columns. These can be either acidic, basic or zwitterionic compounds:
https://helixchrom.com/compounds/23-dih ... zoic-acid/
https://helixchrom.com/compounds/m-toluidine/
https://helixchrom.com/compounds/2-amin ... conditions

Contact me through our website if you have questions.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Thanks Vlad for your for your suggestion. Un fortunately these columns are not with us. I have also performed the exercise on CN column. I am getting much positive results. and getting good separation. Two isomers are overlaying on each other which are very well separated on GC. Three isomers are separated each other. SO for this sample analysis I am performing two exercise. One is GC and other is HPLC.

I required some more and suggestion which gives me very well separation with sharp peak shapes

Thanks
Praveen
Hi Praveen,

I would definitely take Vlad's suggestion more seriously-- separating isomers is a feature specific to specific columns. Separating some isomers will be possible on standard common columns like the CN or C18, but for better resolution a chiral selector is usually the way to go. I've had success separating diastereomers on a PFP(2) column from Phenomenex, but it could not resolve their enantiomeric pairs (4 diastereomers, 4 enantiomers) and would come out as the pairs of diastereomers. The PFP is marketed as a column that has some chiral separation, but clearly not enough in my case. We went through their screening process and found a chiral column that could split these pairs and we're now able to get all of the products efficiently. Positional isomers may be more difficult due to the subtle differences observed between the molecules, which may mean that you have to bump up your chromatography to that next-step. Alternatively, you could purchase (find, if you have some available) some chiral mobile phase additives which is a good alternative to purchasing a new column (depending on intended use of course). It sounds like you have them mostly separated, and depending on if you're running a gradient or not, you may be able to tweak the ratios of your mobile phases slightly (probably extending the overall run-time of your method) in order for your compounds to interact with the column more, and hopefully enhance your resolution. From what you said, it seems like you are able to separate them sometimes, and sometimes they come out as a single peak, this indicates to me that you could probably separate them but it depends on the gradient of your mobile phase.

As always, any conditions, mobile phase details, potential columns, etc. would be very welcome and help us assist you to reach your goals!

Thanks for reading
I second Tyler's opinion.

I have had success with a PFP column before, so maybe try that first. In order to get the three positional isomers separated by HPLC, you might just have to buy a column and Vlad's do appear to be good options.

Read this article: Journal of Liquid Chromatography & Related Technologies 28(19):3131-3142

and this one: Journal of Chromatography A Volume 866, Issue 2, 14 January 2000, Pages 281-2
5 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google Feedfetcher and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry