Ciprofloxacin HPLC/FLU detection

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hello,
I'm using a Shimadzu Prominance LC-20AR HPLC/FLU for ciprofloxacin detection. For stock solution I use Vetranal® ciprofloxacin analytical std. in Methanol (not the HCl form).

My problem is, when I try to measure the calibration points repetitions, the standard deviation of my data is bigger than 15%.
I tried with glass and nylon syringe filter with 2-12-24 ml rinsing volume.

The pH of the stock solution is ~5.6 so 48% of ciprofloxacin is in cationic form and 52% is in Zwitterionic form. My HPLC eluent is pH 2.3 in this pH ciprofloxacin is in cationic form.

(HPLC parameters: A: phosphoric acid pH 2,3, B: ACN; 72:28%; 1ml/min; inj.vol.:20 ul; T:35 °C, detector: ex: 280nm em: 450nm, Teknokroma Brisa C18 column)

Do you have any advice what should I try, to stabilize my measurements, or what could be the problem?

Thank you very much.
Have you tried matching the stock calibration solution matrix to the mobile phase? (Diluting your stock with initial mobile phase)
The past is there to guide us into the future, not to dwell in.
Thank you for your reply.
No, I shouldn't modify the pH of the solutions, because later I will use it to adsorption tests on soils. It would change the adsorption properties.
3 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot], John Guajardo and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry