lmh » Mon Nov 30, 2020 11:44 am
If you're running exactly the same sample on two different methods, and in one you find the analyte, but you don't in the other, then the problem isn't how you prepared the sample, it's the method.
If you're quantifying by LC-MS then ionisation efficiency will depend on what else coelutes with your target compound, but in a typical MRM method you probably have absolutely no idea what's coeluting because your method will only detect the chemicals it's set up to detect - not everything else that happens to be hanging around.
Particularly, if a reverse phase method has poor retention, your target compound may coelute with a whole load of highly polar stuff that's basically unretained. Basically there aren't many alternative explanations if you're running the exact same vial on the exact same system, just with different solvents/gradient, because you know that both methods can detect paracetamol, because you can see it in your standard curve.
I'm guessing the buffer probably gives better retention. Certainly different retention! Good luck...