lmh » Wed Nov 25, 2020 4:20 pm
The problem that is probably bothering them is that if your line passes at any distance significantly above the origin, then you could have small peaks that are visible to the naked eye but don't show up at all because they give negative values. If, for example, your linearity were so poor that the intercept were 10% of the size of the pure compound peak, then you could have individual impurities of up to 10% without them showing up.
10% is obviously ridiculous. You're probably being very professional about this, but somewhere out there is a lab who are doing the same thing, with an intercept that's 1%, or 2%, and who are busy selling products as 98% pure when they have 15 impurities at around the 1% mark by peak-area, and one impurity around 4%.
The final answer is that your method is good provided the line passes close to the origin. There is an LC-GC article somewhere that touches on the situation, but in reverse. The question was "when is it acceptable to force my calibration curve through the origin?" and the answer was "when it passes within a statistically insignificant distance of the origin anyway".
In your situation, the question is how far down the curve can I go before the fact that I did/didn't force to the origin makes a statistically-significant difference to the result? So the way that I'd answer the question is to stress that you have already tested the LOQ. You therefore have a method that is valid down to the LOQ. Your method may fail on impurities that are a long way below the LOQ, but by definition these are not of any practical relevance because in choosing your method, you ensured that it had an adequate LOQ to pick up all relevant impurities. And you are routinely including the LOQ as part of your three-point calibration curve, so your method is valid within the region LOQ < result < 120%, the range of the calibration curve. And that's in ICH Q1(R1).