Accetable retention time drift for system suitability?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
What are acceptable limits for retention drift in LC for system suitability requirements?

I seen some such as 2%,5% and even 10% difference. What is the most common?
No single answer to that. Among other things, it depends on the resolution of your analytes from interferences. As a broad generalization, it should be small enough to prevent mis-identification of peaks.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
For our OQPQ it is NMT 0.3%RSD on six replicates of caffiene.