We use Agilent 1200 gradient HPLC's using RP and various methods using mobile phases that contain buffer. A colleague put it to me that flushing instruments and columns before and after use actually has no effect on chromatography or operation of the system. The procedure for the instruments is to flush each channel with at least 30ml of a wash before using a mobile phase and regardless of whether all lines are used, and to wash the columns with specific wash and storage solutions depending on the samples used.
So is just leaving the mobile phase in the system and columns OK? This person commented that washing a columns does not really do anything, and that for the lines that are not used, what is the point in flushing.
Personally I don't do this as it is contrary to what I have been taught, however I cant really argue with this person, as they maintain they haven't encountered issues so far.
Obviously they seem like durable machines, and can take a lot, but at what point would you start questioning doing this? Or is just a case that flushing systems is not something that requires strict adherence, and if so *how much is too infrequent?
Appreciate any comments, thank you.