Hey everybody!

i have noticed sometimes my UV spectra are a little bit weird. i get negative absorbance and weird shapes.

does any one know what causes this?

Im using agilent DAD 1260 infinity series and Chromeleon software.

i attached 2 pics. They're two peaks far from another but with almost 'identical' spectra by the looks of this ( i know thats not true )

Image

Image

Is it interference or something? or dirty peaks i.e co-elution?

thanks in advance!

kate
Hi Kate,

I was not able to see your images posted. What do you mean by two peaks far from another? RT?

What are your detector settings, scan range, mobile phase?

If you have an agilent controller or Lab advisor you could run a few diagnostics tests, lamp intensity and cell test might be most relevant. How many hours on the lamps? May be worth checking with a new lamp?
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry