Peaks slightly shift by 0.1 min in standard vs sample, Why?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hello,

I am injecting sugar standards and tobacco samples on HPLC-RID. The standard is prepared in pure water and the sample is extract in pure water. When I compare the peaks found in the standard vs sample, there is a slight shift of about 0.1 min in the sample. Why is this even though diluent is the same?
One thing I can think of is tobacco has a lot of salts in it, so maybe that is creating a slight buffer solution with a different pH than the standards and causing the retention time shift.

Have you tried using a buffer solution for both extracts and standards? It might be what is needed to stabilize the retention times.
The past is there to guide us into the future, not to dwell in.
Are the peaks youre comparing the same area?
… and if you need to demonstrate that the shift is caused by the chemistry of the sample, and isn't merely that you're looking at the wrong peak, you can run a mixed vial containing a mix of standard and sample. If the sample is causing the standard to elute at a slightly different retention time, you will get one peak. If the peak in the sample isn't the same as the standard, you will get two peaks.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry