Calibration Question
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
If I analyze for multiple analytes, do all analytes need to be present in all solutions? Can I calibrate 1 analyte using one set of calibration standards and then calibrate a second analyte using a separate set of calibration standards? The assumption here, is that both analytes are quantitated in a single chromatogram with adequate resolution. A source would also be helpful!
-
- Posts: 1861
- Joined: Fri Aug 08, 2008 11:54 am
You can process data where you've got two independent sets of calibration standards containing different calibrants, but it's a real inconvenience and there is no benefit to doing so; it's the sort of thing you'll do if you're running someone else's samples and they've given you the standards ready-prepared but not all in the same vials, or if you realise half way through the batch that you want to add an analyte, so you stick an extra calibration series on the end (hey, bracketing's for wimps...)
The easiest way to do it is just to set up a processing method for standard set A, fill in all the details in your sequence, and process it, then set up a processing method for standard set B, modify your sequence, and process it again. So basically it's double the work. I don't believe there are any specific issues about the validity of this - both halves of the process are completely normal analyses unaffected by the other half.
But of course you're quite right about resolution. You can't do this just because standards A and B run on top of each other and aren't properly resolved, because then they won't be properly resolved in the samples either.
The easiest way to do it is just to set up a processing method for standard set A, fill in all the details in your sequence, and process it, then set up a processing method for standard set B, modify your sequence, and process it again. So basically it's double the work. I don't believe there are any specific issues about the validity of this - both halves of the process are completely normal analyses unaffected by the other half.
But of course you're quite right about resolution. You can't do this just because standards A and B run on top of each other and aren't properly resolved, because then they won't be properly resolved in the samples either.
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google Feedfetcher and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google Feedfetcher and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science