M537 PFC analysis challenges
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
Hi all, I am working on setting up a new EPA Method 537 at our lab for the analysis of perfluorinated compounds. We are having trouble with low recoveries of the lighter/early eluting compounds (i.e. PFBS, PFHxA). Everything recovers well when run as a pure standard, but the problem arises when doing the extraction with actual water samples. Anyone have thoughts on possible causes of the losses? Thanks!
-
- Posts: 617
- Joined: Mon Mar 28, 2005 5:38 pm
The following step is the most likely to cause problems with lighter, more volatile compounds: The extract is concentrated to dryness with nitrogen in a heated water bath,
2 posts
Page 1 of 1
Who is online
In total there are 33 users online :: 2 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am
Users browsing this forum: Bing [Bot], Google [Bot] and 31 guests
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am
Users browsing this forum: Bing [Bot], Google [Bot] and 31 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science, @Food_Science_, @Pharma_Science_
- Follow us on Linkedin: https://www.linkedin.com/showcase/separation-science-/