Started off at 280 bars. I've tried tracing the tubes back sequentially to see if there's a blockage, but that didn't change the pressure.
We initially thought it was the column, but replacing the column with an old one we knew wasn't blocked didn't so anything either - the pressure was still far higher than it should be. We also changed the prefilter in case that was clogged, and also connected in the column without the prefilter, which also didn't affect the pressure.
Overnight we ran an IPA wash through the system, to clear any blockages. This brought the pressure down slightly to around 250 bars, but this is still too high. We tried rinsing the flow lines in case there was a blockage, but this hasn't made any difference.
I've also tried reverse rinsing the outlet tubing with IPA, as it suggests in the autosampler manual, but the pressure is still higher than it should be.
As we still wondered whether it was the column, I reversed and cleaned it, as per the manufacturer's instructions. This brought the pressure down to 190 bars.
Putting a brand new shiny column straight out of the box didn't do anything either. The pressure is still 190 bars. I know new columns can have varying pressures, but that make (Kinetex), the highest a new one has ever been new on our system with that method is 150 bars. But 130 - 140 is about where they normally are when new. 190 is abnormal.
Anyone have any ideas about where to go next? Everything is screaming there's a blockage somewhere, but I cannot for the life of me find it.