vmu » Tue May 05, 2020 9:53 am
There can be virtually no problems with a C18 column in 60:40 methanol:acetone. However, 1) this mixture is a strong eluent, 2) ELSD is a universal detector sensitive to any non-volatile substance and to microparticles (debris) eluting from the column.
A C18 column can contain some hydrolyzed C18 groups (e.g. octadecyl(dimethyl)hydroxysilane), which are hydrophobic and strongly adsorbed. The strong eluent can wash them out to the detector resulting in elevated background signal. Also the problem may be caused by impurities in acetone. Additionally, you can try to filter the mobile phase through a 0.1 μm filter (first check the filter material resistance to the eluent).
If the debris from the column is the cause, it may be necessary to wash the column for a long time to reduce the background or necessary to change the column.