I need help with cleaning. Is there anyway to clean the INSIDE OF THE SYRINGE WITHOUT taking it apart? I understand there is wash solvent and purge solvent. But h class only cleans outside. How do you clean inside of the needle? Without drawing sample and generating data.
-
- Posts: 1
- Joined: Fri Dec 20, 2019 2:21 pm
Prime the syringe from the console.
-
- Posts: 7
- Joined: Thu Nov 15, 2018 1:05 pm
UPLC HClass has a flow trought design, in that way, to clean the inside of the needle you just need to pump the solvent you want and you are good to go.
To clean the seringe you go to UPLC Console>Sample Manager FTN>Control>Prime.
To clean the seringe you go to UPLC Console>Sample Manager FTN>Control>Prime.
3 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science