A few weeks ago I started seeing low intensity (<50 counts) at wavelengths lower than 250 nm. The lamp had 1700 hours of burn time, so that seemed to be the problem. After installing a new lamp (purchased 2 weeks ago from Agilent) and performing calibration, which it passed, the intensity counts at low wavelengths was still essentially 0. I borrowed a new flow cell from a working instrument to see if that was the issue, but once again I'm seeing no intensity at <250 nm.
I'm all out of ideas here. I spoke with Agilent and they think it might be an electronics issue, but that seems weird to me because all other wavelengths look great.
Any other ideas or things I can try?