So, you're looking for horror stories?

I've seen a few.
My list is topped by the ethanol fermentation industry as noted earlier. The standard method uses 0.05 M H2SO4 and an ion exchange column with RI detection (ugly and slow!).
One lab was having continuing column plugging problems (in an earlier visit I had explained the value of filtering, so that wasn't the problem). I ignored the fact that there were huge colonies of fuzzy fungi (or whatever it was) in the waste bottle; this was the least of their problems. Actually, they seemed to be doing most things correct - changing solution regularly, cleaning the bottles between changes, etc., but they still were having problems. But when I looked at the inside of the reservoir cap, there was a ring of black slime. There were bugs growing in/on the cap. Every time they changed the bottle with clean mobile phase, the re-innoculated the solution when they screwed on the cap. Now, the capy was easy to clean, but the LC ...
Another ethanol plant (my first visit), hadn't learned the value of filtering yet. And they seemed to think that the third shift people would find this to be either too difficult or too time-consuming.

The entire instrument was a microbiology experiment!
In both cases I used a 10% bleach solution to clean the lines, the degasser, and the pump. It was a little over the top, but I was so disgusted with how the instrument was being abused that I responded in kind!

When we looked under the microscope at the visible stuff coming out of the degasser, we saw a nice collection of fungal mycelia and other "things." Remarkably, both instruments came back fine; it's a testament to their ruggedness.
Each event only took a few hours to fix, but it was a full day before the systems were up and running again.