LCMS ISTD Area count drops with patient samples
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
5 posts
Page 1 of 1
Lately when I run a batch, my ISTD area count drops immediately with patient samples, but not with calibrators or controls. Some compounds drop progressively during the run and others the drop is drastic. Even when I reinject the calibrators as unknowns (not calibrators) at the end of the batch, they are also affected. Service has been cleaning everything (rods, Q1, Q3, collision cell, optics, etc) with no help. I've ruled out different things but we still can't find what the problem is. Does anyone have any suggestions on what's going on with this instrument?
-
- Posts: 70
- Joined: Fri Jan 02, 2009 12:53 pm
- Location: Los Angeles, CA
from what you describe it appears your sample prep is not very efficient in removing major matrix components
can you elaborate on the nature of patient samples and what is your sample prep and target analytes?
can you elaborate on the nature of patient samples and what is your sample prep and target analytes?
-
- Posts: 2
- Joined: Thu Jun 06, 2024 9:50 pm
- Location: US
We’ve been doing the same sample prep for years and never had this issue before. I even ran the same samples on a different LCMS and the run was fine but in the one at the lab still had the issue.
We process urine samples to detect different kind of narcotics.
We process urine samples to detect different kind of narcotics.
-
- Posts: 2143
- Joined: Tue Aug 17, 2004 7:59 pm
- Location: 38.590967, -90.213236
When cleaning fails, isn't a leak usually the next thing to look for (several times as they can be really hard to find)?
Thanks,
DR

DR

-
- Posts: 70
- Joined: Fri Jan 02, 2009 12:53 pm
- Location: Los Angeles, CA
okay so it should not be that bad indeedWe process urine samples to detect different kind of narcotics.
do you only use positive ionization in your method? how about charging (say, there's electronics issue and one of potentials is off or missing)
if working in positive mode only, you could try adding a dummy transition(s) in negative polarity to flush the charge
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science