Advertisement

Agilent 1290 Infinity II Primary Pumphead Pressure Problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
The primary pumphead on channel B of our Infinity II binary pump will not come up to pressure during the pump leak rate test, it reaches barely 600 bar and stays there for 30 seconds before dropping. I just replaced the pumpseals and rear seal wash seals and everything is tight, no visible leak. I have also tried purging with IPA. What can I do next to troubleshoot?
First, make sure you have fully primed ALL channels with fresh solvent that is degassed (IPS preferred for testing). The pressure leak tests will fail if the system (pump) has trapped gas inside or needs service. This is the number one reason why testing fails.
Is the integrated degasser working? If the degasser is more than 3 years, then it is broken and due for replacement. They do not last long in these units are require regular servicing (not, the degassers are NOT serviced as part of any PM service. They are far too expensive for that and not part of most service plans).

If the system has been serviced, has a working degasser, fully primed, then Replace the pistons too. The pistons and seals are a set and should be replaced if a seal replacement alone does not re-establish function. It should also be stated that you must also be sure to Check the pump's check valves too, esp the outlet valves.
Thanks for your reply. Ultimately we had an Agilent technician come out to check it because absolutely no fault could be found in the pumphead - in the end he exchanged the entire thing saying it was defect.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry