Velos Pro LTQ calibration issue
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
3 posts
Page 1 of 1
We have a Thermo Velos Pro LTQ (not with an orbitrap), we just got it back up and running but have been unable to get it calibrated correctly. The issue we are running into is that the calibration mass at m/z 138 will not pass because it keeps showing up at m/z 139. All other m/z, both lower and higher than 138, pass without a problem. One thing to note is that m/z 138 is created by in-source fragmentation of caffeine at m/z 195. Any suggestions would be greatly appreciated.
-
- Posts: 1862
- Joined: Fri Aug 08, 2008 11:54 am
If you infuse solvent with no calibrant, do you see the mysterious 139? (i.e. is it a contaminant that's got nothing to do with the calibration solution)
If you infuse caffeine, and do in-trap fragmentation on it, do you see 138 or 139? (i.e. is it something where it's got really badly calibrated and it's actually mis-measuring the fragment, or is it that something weird is going on in the spray chamber).
It's a long time since I've used one of these instruments, so I can't remember if it's possible to go back to an old calibration and look at what happens there?
If you infuse caffeine, and do in-trap fragmentation on it, do you see 138 or 139? (i.e. is it something where it's got really badly calibrated and it's actually mis-measuring the fragment, or is it that something weird is going on in the spray chamber).
It's a long time since I've used one of these instruments, so I can't remember if it's possible to go back to an old calibration and look at what happens there?
-
- Posts: 44
- Joined: Tue Oct 04, 2022 10:49 pm
ear0036 wrote:
We have a Thermo Velos Pro LTQ (not with an orbitrap), we just got it back up and running but have been unable to get it calibrated correctly. The issue we are running into is that the calibration mass at m/z 138 will not pass because it keeps showing up at m/z 139. All other m/z, both lower and higher than 138, pass without a problem. One thing to note is that m/z 138 is created by in-source fragmentation of caffeine at m/z 195. Any suggestions would be greatly appreciated.
Is it possible to do a course calibration on your own? Our Finnigan LTQ, while older, allows us to do that.
3 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science