Advertisement

ELSD signal wont show up in chemstation offline analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi everyone

We recently coupled a Sedere LT 90 ELSD to our 1200 Agilent system. We also installed a driver that is compatible with chemstation and it seems to be working fine. The signal (voltage) is shown on the online spectrum in mV and we are able to run and observe the run online. However, when the report is generated and the file is transferred to chemstation offline, only the DAD1 signal is recorded and can be analysed.

Any ideas how to fix this? Greatly appreciated. In signal preferences, I have already added both the ELSD signal and DAD.
I'm retired, but remember a similar situation from way back. We could see such external signal on the monitor but the signal was not being stored. The solution was to run a 9-pin remote cable from the 35900 to the computer.
Thanks, CPG! That was it. Much appreciated.
Thanks, CPG! That was it. Much appreciated.
Winner, winner, chicken dinner !!!!!

The blind squirrel finds a nut every once in a while......
4 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry