Advertisement

strange HPLC mobile phase channels problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi,

I came into the lab this morning and found a very strange problem with the Agilent 1260 Infinity II with a quaternary pump: the water in channel A travelled into channel B. The instrument was in standby mode overnight so the pump didn't operate. How did this happen? Anyone had a smiliar experience?

Regards

Hao
Most likely your Multichannel Gradient Valve is defective.

Read through this tutorial and see if it helps to solve your issue:

https://community.agilent.com/knowledge ... =community
Most likely the seals are damaged or dried out in your MCGV. Make sure you use ALL of the channels in the pump on a regular basis. Do not cap any off or just leave them sitting in unused solutions. After removing all of the lines (4) from the bottles, try flush each one (100% A, etc), one-at-a-time. If you use the system with ANY buffers, start with a 100% pure water flush, then with pure IPA at a normal flow rate for 10-20 minutes EACH to remove deposits and clean the seals.

For more info, please refer to this unbiased article on how to diagnose, troubleshoot and solve the issue.

"HPLC Gradient Valve / Proportioning Valve / MCGV Leaks. How to Identify Them"; https://hplctips.blogspot.com/2021/08/h ... valve.html
Thank you both and I will do the troubleshooting following the instruction.
4 posts Page 1 of 1

Who is online

In total there are 6 users online :: 2 registered, 0 hidden and 4 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot], Bing [Bot] and 4 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry