If the UV spectra for the additives you're looking for have any distinct local maxima, you might be able to ditch HPLC in favor of a UV/Vis method instead.
Lots less maintenance, less time checking samples, far fewer chemicals to buy... whether this would work is a function of whether the other mixture components absorb too strongly at a given wavelength.
I spent a decent amount of time yesterday running UV/Vis scans on three concentrations of the additive; 0.5%, 1.0%, and 3.0%bv. I ran multiple modes with rather poor results. Using absorbance, the three peak at slightly different wavelengths; 295, 297, and 301nm respectively. When comparing the absorbance ratios at their peaks, they are nowhere near the ratio of conc., and of course less accurate if selecting 297nm as that maximizes the 1.0% sample, but then hits off-peak on the other two. I did not bother integrating the areas since the peaks do not coincide.
UV/Vis is a Thermo Genesys10s using the VisionLite software.
Well, now you know. I guess those maxima are way too close together to be useful.
Thanks,
DR
Image
17 posts Page 2 of 2

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry