This may be due to the recent samples we've been getting (despite the fact sample prep involves a 0.22 um filtration), but we've been having heavy peak tailing issues on our Agilent HPLC system.

We would flush the column with mobile phase (we run an isocratic run with 0.005 N sulfuric acid on a Supelcogel C-610H 9 um 30x7.8 cm column) - reverse the column and the guard column so the flow is in the opposite direction - and it would fix the problem at first. But after even a few sample injections the tailing would appear again.

I realize the dirty samples might not be helping, but any advice to keep the tailing at bay for longer? Such as good flush solvents or guard columns to help keep the system clean enough? We just bought the column 3 months ago.

Thanks so much!
I would speculate that the guard cartridge (or possibly, but less likely the column) has a head space. When you reverse the flow, it eventually moves to the other end. If this were my problem, I would start by changing the guard cartridge.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Completely agree with Tom as a void at the head of the column is the most likely (and in fact very common) reason. These gel columns are sometimes quickly compressed, fouled and/or damaged. It is not your HPLC system's fault. Columns are consumable items, meant to be replaced. If the column is not lasting very long, then review your method to insure it follows good chromatography principles, do not backflush gel columns and of course review your sample prep procedures. Have someone with HPLC experience review your method and techniques on site to find the true cause and make suggestions or provide training to solve.
3 posts Page 1 of 1

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry