Carryover peak RP-HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
I work on RP-HPLC method for analysing a small peptide. My diluent is water. Mobile phase A is Water +0.065% TFA & Mobile phase B Acetonitrile +0.05% TFA. I use Inertsil OTS-3 column together with guard column. When the peptide is mixed with CMC (Carboxymethyl cellulose that is natural glue) and mannitol I have massive carry over which I believe is due to the presence of CMC. It takes about 5 water injections to remove the carryover peak. Do you have any ideas for removing the carryover? I have eliminated other obvious contamination factors and I came to conclusion that this must be column retaining the sample that did not elute in the first run. Thank you in advance for your help.
The carboxyl- groups of the polymer may be chelating heavy metals in the HPLC column. Try passivating the column by pumping a solution of 40 mM EDTA.2Na through it at a low flow rate for at least 16 hours. Then, flush out the EDTA, equilibrate with your mobile phase, and assess the degree of carryover that you get. If this solves the problem, then repeat this treatment every 3 weeks. If it improves the situation but doesn't solve it, then try treating with EDTA for 24-30 hours instead of 16 hours.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
Hello

Few ideas:
1. Extend final holdup time (high ACN %) in your method and check few blanks after to confirm if all "rubish" is eluted from column
2. You can create method with 95:5 ACN/Buffer and run it after your sample to clean column. Then you can run blank/no injection to confirm there is no carryover
3. Don't forget to setup proper needle/needle seat washing for sampler - it is critical place for carryover issue.

Regards

Tomasz Kubowicz
Hello

Few ideas:
1. Extend final holdup time (high ACN %) in your method and check few blanks after to confirm if all "rubish" is eluted from column
2. You can create method with 95:5 ACN/Buffer and run it after your sample to clean column. Then you can run blank/no injection to confirm there is no carryover
3. Don't forget to setup proper needle/needle seat washing for sampler - it is critical place for carryover issue.

Regards

Tomasz Kubowicz
4 posts Page 1 of 1

Who is online

In total there are 34 users online :: 0 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry