Glyphosate is not reproducible

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

12 posts Page 1 of 1
Hi,

I am having trouble with analyzing and developing a method for glyphosate.

I had the method working and could see glyphosate down to 0.01mg/kg and then all of a sudden the peak deteriorated and I haven't been able to get it back.

I am using a Shimadzu LC stack with the mass spec; Sciex API 4000. An Acclaim Trinity Q1 column 3um 3x100mm (new column)

Mobile phase A: 50mM ammonium formate, formic acid 2.9pH, H2O:ACN 6:4, Mobile phase B:ACN

Gradient Binary Flow: 0 mins 100% A, 10.1 mins 10% A, 13.1 mins 100% A, 18 mins 100% A

Flow rate of 0.5ml/min

(this is from the EU reference glyphosate method: http://www.crl-pesticides.eu/userfiles/ ... urlSRM.pdf)

I prep the samples using the QuPPe method therefore, it is not derivatized.

Since the problem arose, I have made fresh mobile phases and cleaned the ion source. I don't think it's a detection problem as the detector works on other methods.

Does anyone have any idea why it was working and now it is not and how to get it working again? Any help would be appreciated.
Do you see it at higher concentration or you don't see glyphosate at all? Are your injections from the same vial or you made new samples? Did you try to inject glyphosate without the column to see if it related to column or HPLC system?
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Thanks so much for replying, this glyphosate is driving me slightly mad!

When the samples are being extracted there is barely a peak and very noisy at 2mg/kg. When I run a solvent standard at 2mg/kg I am getting a better peak but the sensitivity is still terrible and isn't anywhere near as close to where I had it last week with the sensitivity for an extracted 2mg/kg sample.

I have tried lots of new extractions, tried some from old vials in case it was a new extraction but they all seem to give the same terrible noisy peaks.

I've injected the solvent glyphosate std without a column and it is present, when I tried taking the column out the ferrule was really stuck in the column and was a hard job to get out. I'm just about to change that and see if the problem is just as simple as having to change a ferrule.

Edit: Unfortunately it was not as simple as changing the ferrule. To give you an idea of what I had last week compared to now I'll try and attach the images below. Image
is it the same response ad with the column? If it is more without column, then you column traps glyphosate or noise level from the column reduces you sensitivity.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Those are both with the column, the same column. The one on the left was from last week when the method was working and the other is from today. The one on the left is actually only 0.6mg/kg and the one on the right is 2mg/kg.
I agree with Vlad, it seems a good test to replace the column with a zero-volume connection and do the same injection of a solvent standard. Don't forget to adjust the MS method for this test: no divert valve, and your analyte will obviously elute earlier.

On a sidenote, I was interested in your column, googled it and I have to say it's quite a special one (for example, it can not be used with MeOH). Your mobile phase choice seems to be compatible, altough the pH might be a bit close to the lower limit. Let's see what the no-column test gives?
The link below might be interesting for your issue:

http://www.waters.com/webassets/cms/eve ... ebcast.pdf

In slide 31 they mention strong peak tailing possible due to metal ion contamination which they could mitigate with an acid wash. No more details though,.
So without the column the intensity of the solvent peak is lower than with the column.

Thanks for that link, it looks as though that wash could be of help, very annoying they don't say what sort of acid wash they use, that would've been very handy. I had a google and the Acclaim trinity manual has its own suggestions of washing the column so I will give that a go. https://assets.thermofisher.com/TFS-Assets/CMD/manuals/Man-065475-Acclaim-Trinity-Q1-Man065475-EN.pdf

It is a special column. From doing research it seemed that this column gave the best results for glyphosate so we purchased it hoping it would be a case of using the mobile phases that it suggests along with the same gradient and doing the QuPPe extraction. It did work very well for about a week and then the peak just started to get worse and become inconsistent and not usable. I think that's the most frustrating part, I had it working.

It could be a case that the column needs to be maintained regularly in order to give good chromatography. I shall see what happens after doing the washes they suggest.
Possibly your column is interacting with the phosphonate group. Passivate the metal surfaces of your column (as well as the rest of the HPLC system) by eluting it overnight at a low flow rate with 40 mM EDTA.2Na. Next morning, flush out the EDTA, condition your column, and evaluate the recovery of injected glycosate.
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
In the past we used a Hamilton column for Glyphosate along with one from Pickering and both needed to be cleaned at times due to iron contamination. Usually a few injections of 100ul of about a 5% Acetic acid solution would do the trick. Have to see if that would be compatible with this column you are using.
The past is there to guide us into the future, not to dwell in.
Thanks for the replies. Cleaning the column over night did not seem to help. I'm having to resort to derivatization and hopefully I can now get consistent results.

Currently prepping the samples now so I will have to wait and see.

Thanks for all the help :)
PaulaL wrote:
So without the column the intensity of the solvent peak is lower than with the column.

It could be a case that the column needs to be maintained regularly in order to give good chromatography. I shall see what happens after doing the washes they suggest.


This means that the column is not the problem, or do i misinterpret the line in bold above?
12 posts Page 1 of 1

Who is online

In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 12 guests