Hi.
I have been working with Aminex HPX87H columns for fermentation analysis for very long time, using photodiode array (PDA) and refraction index detectors, with H2SO4 as mobile phase.
In order to transfer the column method to a system incorporating PDA and QDa detectors (Waters, MS, electrospray), I wonder if it is possible to replace the mobile phase with some other acid, more compatible with this detector, since sulphuric causes ion supression. I thought of formic acid, but I don't know if it affects the resin somehow or saturates the column.
Any thoughts?
Ion suppression in the mobile phase is the whole point behind using sulfuric acid. What you want is a strong acid that is not very hydrophobic (so it doesn't stick to the resin), which pretty much limits you to mineral acids.

You might try TFA, but I'd advise doing on an expendable column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry