Why not consider the 1.7 µm columns? If you can get the peak witdh down to seconds you will have a decent chance to get at least partial separation. You will prefarably need a UPLC, but you can get quite far with a standard LC (<400 bar) with optimised components (minimised dwell volumes, small volume flow cell).

With a 100 mm column you can typically use 0.4 ml/min without exceeding the pressure maximum.

Ha,ha, you are talking 2 dimentional HPLC

Tons of research articles, but not easy to apply in real life

see this attached paper, it tells the major technical difficulties

http://web.ebscohost.com/ehost/pdf?vid= ... 4%40SRCSM2

firework1, true, some are difficult to set up, but once you get it going it works when all attempts at single column or series columns fail. A real life example: J Chrom B, 678, 137 (1996).
Also, this is not really rare at all, as all methods using SPE for cleanup are two step chroms (two dimensional).

Thanks for all of your comments.

Due to a serious lack of time lately to really get into the meat of such a separation during the day (I have validation protocols to write), I've been running my AMDS nightly on a series of samples containing mixtures of 6-8 or so analytes grouped roughly by retention time. The composition of each sample overlaps each "adjacent" one by at least 2 compounds on both ends. For instance, sample one contains the first 6 eluters, sample two contains 4-12, sample three contains 10-18, and so forth.

I'm using alot of solvent, but thus far results appear to be promising and it's automated, so what the heck.

With care and some tweaking, I hope to be able to string optimized gradients together, check the effects of column dimensions / particle size for further optimization, and come up with a reasonable method on a single column.

So far I have 13 well defined peaks (min Rs ca 1.7), except for the cis-trans isomers of one, with which I'm not terribly concerned.

If it works, I'll post a chromatogram. If it doesn't, I'll go the two column route...and maybe try the whole thing over again that way.
19 posts Page 2 of 2

Who is online

In total there are 15 users online :: 0 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry