Hello, i launched the USP technic dedicated to Azithromycin tablets organic impurities dosage. The System suitability solution were injected six times , unfortunately the RSD for Area were higher than the 2.0% mentionned (3.6% for azithromycin related compound F , and 13.4% for desosaminylazithromycin) . The case is that the fifth injection were shifted to the right and weren't stackable with the others . Any suspected reason for that ?
I can think of many possible reasons, but the fact that you saw a retention time change on one of the injections suggests a flow rate problem. One possibility is that the one bad injection is the cause of the high RSD (what does the RSD look like for the remaining 5 injections?).
Also, is there a trend in the response (in other words, is the high RSD coming from a gradually increasing or decreasing response rather than random variation)?
Have you seen acceptable RSD with other methods run on this system ?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Thank you for answering .
The retention time increased from injection 1 to injection 6 , and the RSD for this method is mentionned in the monograph =< 2%.

Image
This is the summary of the technic results

Image
And this is the graph of the sixth injection .
Unfortunately, the images you linked to are not accessible. Error message was:
"403. That’s an error.
Your client does not have permission to get URL
"
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Sorry it was my mistake , here are links to access to the pictures on google drive

https://drive.google.com/open?id=0BwJxH ... 01YWhueFB3


https://drive.google.com/open?id=0BwJxH ... xfRjVxU19n

and sorry again for being late
Hello. As far as I see it seems to be an integration issue. Try to tweak integration options and reprocess your data or try manual integration.
Best regards,
Dmitriy A. Perlow
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry