Hi,

Does anyone know a free software that can calculate the solvent needed for a gradient in HPLC?

Thank you!

GB.

Use Excel.
If you're doing a linear gradient between 2 different %B, for exaample, multiply the average %B at the start/end of the gradient step times the total flow rate to get the mL B for that step.
Repeat for A.
Repeat all of the above for however many steps you have.
Sum that mess for ea. A & B to get a total mL of ea. per injection.
Multiply ea. sum by the number of injections to get totals (remember to include however long your autosampler turn-around time is as an addition to your initial conditions) for mL A, B for your run (then make a little extra).
Thanks,
DR
Image

You can write a spreadsheet to do it. If it has to be free, get Open Office.
Mark Tracy
Senior Chemist
Dionex Corp.
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 0 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: No registered users and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry