Having trouble figuring out the cause of my rolling baseline. The roller is very symmetrical and has a peak to valley time of approx. 10 minutes. Won't bore you w/ the details of my trouble shooting. Any thoughts?
s

When the time between two injections correlates with the baseline rolling time you see a (very) late eluating peak.

the rolling baseline occurs during equilibration as well as during runs.

How about temperature fluctuations? Air conditioning going off and on? Column heater off/on?

If the problem is caused by a late eluating peak, you get rid off the problem by flushing the column with high % to 100% organic solvent.
The next equilibrating with your normal mobile phase should show a smooth baseline.
Using a solvent with 10 to 20% less water show such a compound coming from the sample(s) within normal run time.
5 posts Page 1 of 1

Who is online

In total there are 15 users online :: 1 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry